Irrigated Bipolar Radiofrequency Ablation with Transmurality Feedback for the Surgical Cox-Maze Procedure

Authors

  • Chad E. Hamner
  • Alison Lutterman
  • D. Dean Potter
  • Thoralf M. Sundt III
  • Hartzell V Schaff
  • David Francischelli

DOI:

https://doi.org/10.1532/hsf.733

Abstract

Background: Nonirrigated radiofrequency ablation (RFA) has been used to replicate the surgical scars of the Cox-Maze procedure. This study aimed to demonstrate that an irrigated, bipolar RFA energy source could also effectively replicate Cox-Maze lesions with impedance monitoring to predict the transmurality of ablated tissue.

Methods: A complete Cox-Maze lesion pattern was created ex vivo on fresh porcine atria using an irrigated, bipolar RFA system. Tissues were clamped between opposing electrodes with steady pressure to ensure an intimate tissue-electrode interface during ablation. A proprietary feedback and control algorithm monitored tissue impedance and terminated ablation when lesions were deemed transmural by a plateau in impedance decline. Ablation time and power, lesion width and length, and tissue thickness were recorded. Lesions were stained with 1% triphenyltetrazolium chloride and sectioned for gross assessment of transmurality.

Results: One hundred thirty-seven lesions were created on 11 porcine hearts. The total ablation time per lesion was 14.8 ± 1.2 seconds (range, 10.0-19.0 seconds). Lesions averaged 4.2 ± 1.3 mm (range, 1.3-10.2 mm) in width. Average tissue thickness was 3.0 ± 1.7 mm (range, 0.5-9.9 mm). Cross-sectional examination revealed that 100% of lesions were transmural (n = 718), and no tissue defects were observed.

Conclusions: These results indicate that irrigated bipolar RFA energy can produce transmural Cox-Maze lesions ex vivo on intact porcine atria and that impedance monitoring is a reliable predictor of lesion transmurality. Additional in vivo studies are under way to further demonstrate the efficacy and safety of irrigated, bipolar RFA technology.

References

Arcidi JM Jr, Doty DB, Millar RC. 2000. The Maze procedure: the LDS Hospital experience. Semin Thorac Cardiovasc Surg 12:38-43.nBonanomi G, Zenati M, Francischelli D, et al. 2002. A new device for intraoperative left atrial ablation [abstract]. Circulation 106:II-669.nCaccitolo JA, Stulak JM, Schaff HV, Francischelli D, Jensen DN, Mehra R. 2001. Open-heart endocardial radiofrequency ablation: an alternative to incisions in Maze surgery. J Surg Res 97:27-33.nCox JM, Ad N, Palazzo T, et al. 2000. Current status of the maze procedure for the treatment of atrial fibrillation. Semin Thorac Cardiovasc Surg 12:15-9.nDemazumder D, Mirotznik MS, Schwartzman D. 2001. Biophysics of radiofrequency ablation using an irrigated electrode. J Interv Card Electrophysiol 5:377-89.nDemazumder D, Mirotznik MS, Schwartzman D. 2001. Comparison of irrigated electrode designs for radiofrequency ablation of myocardium. J Interv Card Electrophysiol 5:391-400.nDoll N, Borger MA, Fabricius A, et al. 2003. Esophageal perforation during left atrial radiofrequency ablation: is the risk too high? J Thorac Cardiovasc Surg 125:836-42.nFishbein MC, Meerbaum S, Rit J, et al. 1981. Early phase acute myocardial infarct size quantification: validation of the triphenyltetrazolium chloride tissue enzyme staining technique. Am Heart J 101:593-600.nGaynor SL, Ishii Y, Diodato MD, et al. 2003. A complete Cox-maze procedure performed on the beating heart with bipolar radiofrequency energy: operative technique and early results. Heart Surg Forum 6(suppl 1):S20.nGillinov MA, Pettersson G, Rice TW. 2001. Esophageal injury during radiofrequency ablation for atrial fibrillation. J Thorac Cardiovasc Surg 122:1239-40.nHaines DE, Verow AF 1990. Observations on electrode-tissue interface temperature and effect on electrical impedance during radiofrequency ablation on ventricular myocardium. Circulation 82:1034-8.nMelo J, Adragao P, Neves J, et al. 2000. Endocardial and epicardial radiofrequency ablation in the treatment of atrial fibrillation with a new intra-operative device. Eur J Cardiothorac Surg 18:182-6.nPrasad SM, Maniar HS, Moustakidis P, Schuessler RB, Damiano RJ Jr. 2002. Epicardial ablation on the beating heart: progress towards an off-pump maze procedure. Heart Surg Forum 5:100-4.nPrasad SM, Maniar HS, Schuessler RB, Damiano RJ Jr. 2002. Chronic transmural atrial ablation by using bipolar radiofrequency energy on the beating heart. J Thorac Cardiovasc Surg 124:708-13.nSchaff HV, Dearani JA, Daly RC, Orszulak TA, Danielson GK. 2000. Cox-Maze procedure for atrial fibrillation: Mayo Clinic experience. Semin Thorac Cardiovasc Surg 12:30-7.nSie HT, Beukema WP, Misier ARR, et al. 1997. Radiofrequency ablation of atrial fibrillation in patients undergoing valve surgery. Circulation 96(suppl):450-1.nSonmez B, Demirsoy E, Yagan N, et al. 2003. A fatal complication due to radiofrequency ablation for atrial fibrillation: atrio-esophageal fistula. Ann Thorac Surg 76:281-3.nThomas SP, Guy DJ, Boyd AC, Eipper VE, Ross DL, Chard RB. 2003. Comparison of epicardial and endocardial linear ablation using handheld probes. Ann Thorac Surg 75:543-8.nThomas SP, Nunn GR, Nicholson I, et al. 1997. An endocardial technique for cure of atrial fibrillation during cardiac surgery. Circulation 96(suppl):450.nVivaldi MT, Kloner RA, Schoen FJ. 1985. Triphenyltetrazolium staining of irreversible ischemic injury following coronary artery occlusion in rats. Am J Pathol 121:522-30.n