Early Experience with a New Aortic Clamping System Designed for Port Access Cardiac Surgery: The PortaClamp

Authors

  • Didier de Cannière
  • Mohammad Dindar
  • Constantin Stefanidis
  • Olivier Jegaden
  • Jean-Luc Jansens

DOI:

https://doi.org/10.1532/HSF98.20041012

Abstract

Background: We report a clinical study to demonstrate the feasibility and safety of a new aortic crossclamping concept for use in port access cardiac surgery. The limited access to the aorta in minimally invasive cardiac surgery mandates specific clamping modalities, which entail specific limitations, drawbacks, and costs. Therefore a new autoguided, extravascular, and atraumatic clamping system (PortaClamp) was developed to facilitate port access surgery while potentially avoiding the complications and costs inherent to endoluminal clamping or "blind" crossclamping.

Methods: Twenty patients underwent various cardiac operations under cardiopulmonary bypass and aortic cross-clamping with the PortaClamp between February and September 2003. The method of aortic clamping is described and the operative course and clinical outcome of the patients are reported as surrogates of feasibility and safety.

Results: The average time to position the clamp was 196 ± 75 seconds. Crossclamping through a 10-mm port or incision was achieved successfully, enabling cardiac arrest throughout the procedure in every patient. No patient presented with cardiovascular accident or transient ischemic attack, aortic dissection, or hematoma. Intensive care unit times were 12 ± 3 hours; length of stay was 7.2 ± 1.1 days.

Conclusion: From this early experience we conclude that the PortaClamp system is safe and can effectively be used to crossclamp the aorta inexpensively to facilitate port access cardiac surgery. Further comparative studies with the existing systems are warranted to confirm that the atraumatic design provides further benefit.

References

Hesselvik JF, Ortega RA, Treanor P, Shemin RJ. 1999. Intraoperative rupture of the endoaortic clamp balloon in a patient undergoing port-access mitral valve repair. J Cardiothorac Vasc Anesth 13(4):462-5.nMohr F, Falk V, Diegeler A, Walther T, van Son, Autsschbac R. 1998. Minimally invasive port-access mitral valve surgery. J Thorac Cardiovasc Surg 115(3):567-74.nSchneider F, Falk V, Walther T, Mohr FW. 1998. Control of endoaortic clamp position during Port-Access mitral valve operations using tanscranial Dopplerechography. Ann Thorac Surg 65(5): 1481-2.nVanermen H, Farhat F, Wellens F, et al. 2000. Minimally invasive video-assisted mitral valve surgery: from Port-Access towards a totally endoscopic procedure. J Card Surg 15(1):51.nWimmer-Greinecker G, Matheis G, Dogan S, et al. 1999. Complications of port-access cardiac surgery. J Card Surg 14(4):240-5.nAybek T, Dogan S, Wimmer-Greinecker G, et al. 2000. The micromitral operation comparing the Port-Access technique and the transthoracic clamp technique. J Card Surg 15:76-81.nBarbut D, Hinton RB, Szatrowski TP, et al. 1994. Cerebral embolidetected during bypass surgery are associated with clamp removal. Stroke 25(12):2398-40.nCasselman FP, Van Slycke S, Wellens F, et al. 2003. Mitral valve surgery can now routinely be performed endoscopically. Circulation 108(suppl 1):II48-54.nFelger JE, Chitwood WR Jr, Nifong LW, Holbert D. 2001. Evolution of mitral valve surgery: toward a totally endoscopic approach. Ann Thorac Surg 72(4):1203-8.nGersak B, Sostaric M, Kalisnik JM. 2003. Endoscopic aortic valve replacement. Heart Surg Forum 6(6):E197-9.nGrocott HP, Smith MS, Glower DD, Clements FM. 1998. Endovascular aortic balloon clamp malposition during minimally invasive cardiac surgery: detection by transcranial Doppler monitoring. Anesthesiology 88(5):1396-9.nHagl C, Ergin MA, Galla JD, et al. 2000. Delayed chronic type A dissection following CABG: implications for evolving techniques of revascularization. Card Surg 15(5):362-7.n

Published

2005-01-04

How to Cite

Cannière, D. de, Dindar, M., Stefanidis, C., Jegaden, O., & Jansens, J.-L. (2005). Early Experience with a New Aortic Clamping System Designed for Port Access Cardiac Surgery: The PortaClamp. The Heart Surgery Forum, 7(3), E240-E244. https://doi.org/10.1532/HSF98.20041012

Issue

Section

Article