Pre-Clinical Validation of a New Intra-Operative "Dual Beam Doppler" Blood Flowmeter in an Artificial Circuit
DOI:
https://doi.org/10.1532/HSF98.20051031Abstract
Background: Intra-operative flow measurement during coronary or peripheral bypass operations is helpful for ruling out technical failures and for prediction of complication and patency rates. Preclinical validation of the flowmeters is required in order to rely on the intra-operatively measured results. The aim of this study is to evaluate a new "dual beam Doppler" blood flowmeter before clinical application and to compare it with the established "transit time flow measure-ment" technique in an artificial circuit.
Methods: Measurements were performed in an experimental flow model using pig blood and pig arteries. Three different flowmeters were used: Quantix OR (dual beam doppler flowmeter), CardioMed (transit time flowmeter), and Transonic (transit time flowmeter). Three validation tests were performed to assess correlation, precision, and repeatability of devices. (1) Correlation and agreement analysis was performed with various flow amounts (10-350 mL/min) (n = 160). (2) Device reproducibility and measurement stability were tested with a constant flow (flow amount = 300 mL/min) (n = 30). (3) A user accuracy test (intra- and inter-observer variability) was performed by 5 different observers with a constant flow (flow amount = 205 mL/min) (n = 75). Time collected true flow was used as a reference method in all steps and all tests were performed in a blind manner. Results are shown as mean values ± standard deviations. Pear-son's correlation and Bland-Altman plot analyses were used to compare measurements. Results: The mean flow was 167 ± 98 mL/min for true flow and 162 ± 94 mL/min, 165 ± 94 mL/min, and 166 ± 100 mL/min for Quantix OR, CardioMed, and Transonic, respectively. Correlation coefficients between Quantix OR, Medi-Stim, Transonic, and time collected true flow were over 0.98 (P = .01). Most of the measured results ( > 90%) were between ± 1.96 SD agreement limits in Bland and Altman plot analysis. All devices showed good results in the reproducibility test. During the user accuracy test, larger variance changes were observed between intra- and inter-observer results with the dual beam Doppler flowmeter compared to the 2 used transit time flowmeters when used for single sided vessel access without stabilization device (available from the manufacturer).
Conclusion: All 3 tested flowmeters showed an excellent correlation to the true flow in an artificial circuit and the accuracy of the tested devices was within agreement limits. Reproducibility of all devices was good and linear. The new dual beam Doppler flow measurement technique compares favorably to the classic transit time method. Clinical use may depend on operator, location, and condition, thus more studies may be required to ensure uniform results using the currently available blood flow measurement devices.
References
Alback A, Roth WD, Ihlberg L, Biancari F, Lepontalo M. 2000. Preoper ative angiographic score and intraoperative flow as predictors of the mid term patency of infrapopliteal bypass grafts. Eur J Vasc Endovasc Surg 20:447-53.nBauer SF, Bauer K, Rosendahl UP, Dalladaku F, Ennker IC, Ennker J. 2002. Intraoperative bypass flow measurement reduces the incidence of postoperative ventricular fibrilation and myocardial infarction after coronary artery revascularization. Thorac Cardiovasc Surg 50:S27.nAlback A, Makisola H, Nordin A, Lepontalo M. 1996. Validity and reproducibility of transit time ?owmetry. Ann Chir Gynaecol 85:325-31.nBland JM, Altman DG. 1986. Statistical methods for assessing agreement between two methods of clinical measurements. Lancet 1:307-10.nBeldi G, Bosshard A, Hess OM, Althaus U, Walpoth BH. 2000. Transit time flow measurement: experimental validation and comparison of three different systems. Ann Thorac Surg 70:212-7.nLundell A, Bergqvist D. 1993. Prediction of early graft occlusion in femoropopliteal and femorodistal reconstruction by measurement of volume flow with a transit time flowmeter and calculation of peripheral resistance. Eur J Vasc Surg 7:704-8.nBonatti J, Danzmayr M, Schahner T, Friedrich G. 2003. angiography for quality control in MIDCAP and OPCAB. Eur J Cardiothorac Surg 24:647-9.nCikirikcioglu M, Kalangos A, Walpoth BH. 2005. Intraoperative verification of graft patency in coronary bypass operations with blood flow measurement. In: Textbook of Cardiovascular Surgery, Duran E, ed. (in Turkish). Capa Tip Kitabevi, Istanbul: 939-48.nD'Ancona G, Karamanoukian HL, Ricci M, Schmid S, Bergsland J, Salerno TA. 2000. Graft revision after transit time flow measurement in off-pump coronary artery bypass grafting. Eur J Cardiothorac Surg 17:287-93.nDrost CJ. 2002. Flow based intraoperative coronary graft patency assessment. Transonic Systems Inc.:17.nHarder Y, Leiser A, Canova C, Furrer M. 2002. Transit time flowmea-suring (TTF) in infrainguinal bypass surgery: helpful tool for on table quality control and outcome indicator. Thorac Cardiovasc Surg 50:S38.nJohnson BL, Bandyk DF, Back MR, Avino AJ, Roth SM. 2000. Intraop-erative duplex monitoring of infrainguinal vein bypass prodedures. J Vasc Surg 31:678-90.nKaramanoukian HL, Donias HW. 2003. Intraoperative graft patency verification during on- and off-pump coronary bypass surgery. Medistim AS:5-6.nMuratori M, Berti M, Doria E, et al. 2001. Transesophageal echocardio-graphy as predictor of mitral valve. J Heart Valve Dis 10:65-71.nStirnemann P, Ris HB, Do D, Hamerli R. 1994. Intraoperative flow measurement of distal runoff: a valid predictor of outcome of infrain- guinal bypass surgery. Eur J Surg 160:431-6.nWalpoth BH, Bosshard A, Kipfer B, Berdat PA, Althaus U, Carrel T. 1998. Failed coronary artery bypass anastomosis detected by intraopera- tive coronary flow measurement. Eur J Cardiothrorac Surg 14:S76-81.nWalpoth BH, Bosshard A, Genyk I, et al. 1998. Transit time flow measurement for detection of early graft failure during myocardial revascularization. Ann Thorac Surg 66:1097-1100.nHol PK, Fosse E, Lundblad R, et al. 2002. The importance of intraoper-ative angiographic findings for predicting long-term patency in coronary artery bypass operations. Ann Thorac Surg 73:813-8.nSchmitz C, Ascraf O, Schiller W, et al. 2003. Transit time flow measurement in on-pump and off-pump coronary artery surgery. J Thorac Car-diovasc Surg 126:645-50.nSkladany M, Vilkerson D, Lyons D, Chilipka T, Delamere M, Hollier LH. 1998. New, angle-independent, low cost doppler system to measure blood flow. Am J Surg 176:179-82.nSoustiel JF, Levy E, Zaaroor M, Bibi R, Lukaschuk S, Manor D. 2002. A new angle-independent Doppler ultrasonic device for assessment of blood flow volume in the extracranial internal carotid artery. J Ultrasound Med 21:1405-12.nWalsh D. 2000. Technical adequacy and graft thrombosis. In: Rutherford RB, ed. Vascular Surgery. Philadelphia, PA: WB Saunders Company, 708-26.nLundell A, Bergqvist D, Mattsson E, Nilsson B. 1993. Volume blood flow measurement with a transit time flowmeter: an in vivo and in vitro variability and validation study. Clin Physiol 13:547-57.nRasmussen TE, Panneton JM, Kalra M, et al. 2003. Intraoperative use of a new angle-independent Doppler system to measure arterial velocities after carotid endarterectomy. J Vasc Surg 37:374-80.nRicci M, Karamanoukian HL, Salerno TA, Dancona G, Bergsland J. 1999. Role of coronary graft flow measurement during reoperations for early graft failure after off-pump coronary revascularization. J Card Surg 14:342-7.n