Objective: The use of homografts for aortic valve replacement (AVR) is an alternative to mechanical or biological valve prostheses, especially in younger patients. This retrospective comparative study evaluated our single-center long-term results, with a focus on the different origins of the homografts.
Methods: Since 1992, 366 adult patients have undergone AVR with homografts at our center. We compared 320 homografts of aortic origin and 46 homografts of pulmonary origin. The grafts were implanted via either a subcoronary technique or the root replacement technique. We performed a multivariate analysis to identify independent factors that influence survival. Freedom from reintervention and survival rates were calculated as cumulative events according to the Kaplan-Meier method, and differences were tested with the log-rank test.
Results: Overall mortality within 1 year was 6.5% (21/320) in the aortic graft group and 17.4% (8/46) in the pulmonary graft group. In the pulmonary graft group, 4 patients died from valve-related complications, 1 patient died after additional heterotopic heart transplantation, and 1 patient who entered with a primary higher risk died from a prosthesis infection. Two patients died from non-valve-related causes. During the long-term follow-up, the 15-year survival rate was 79.9% for patients in the aortic graft group and 68.7% for patients in the pulmonary graft group (P = .049). The rate of freedom from reoperation was 77.7% in the aortic graft group and 57.4% in the pulmonary graft group (P < .001). The reasons for homograft explantation were graft infections (aortic graft group, 5.0%; pulmonary graft group, 6.5%) and degeneration (aortic graft group, 7.5%; pulmonary graft group, 32.6%).
Conclusion: Our study demonstrated superior rates of survival and freedom from reintervention after AVR with aortic homografts. Implantation with a pulmonary graft was associated with a higher risk of redo surgery, owing to earlier degenerative alterations.
References
Barratt-Boyes BG, Roche AH, Brandt PW, Smith JC, Lowe JB. 1969. Aortic homograft valve replacement. A long-term follow-up of an initial series of 101 patient. Circulation 40:763-75.nBorger MA, Ivanov J, Armstrong S, Christie-Hrybinsky D, Feindel CM, David TE. 2006. Twenty-year results of the Hancock II bioprosthesis. J Heart Valve Dis 15:49-56.nButany J, Ahluwalia MS, Nair V, David TE. 2004. Cryopreserved pulmonary homograft: postimplant changes. Cardiovasc Pathol 13:59-61.nEdmunds LH Jr, Clark RE, Cohn LH, Grunkemeier GL, Miller DC, Weisel RD. 1996. Guidelines for reporting morbidity and mortality after cardiac valvular operations. The American Association for Thoracic Surgery, Ad Hoc Liaison Committees for Standardizing Definitions of Prosthetic Heart Valve Morbidity. Ann Thorac Surg 62:932-5.nFischlein T, Schütz A, Haushofer M, et al. Immunologic reaction and viability of cryopreserved homografts. Ann Thorac Surg 1995;60:122-6.nFrater RW, Furlong P, Cosgrove DM, et al. 1998. Long-term durability and patient functional status of the Carpentier-Edwards Perimount pericardial bioprosthesis in the aortic position. J Heart Valve Dis 7:48-53.nGorczynski A, Trenkner M, Anisimowicz L, et al. 1982. Biomechanics of the pulmonary autograft valve in the aortic position. Thorax 37:535-9.nGulbins H, Kilian E, Roth S, Uhlig A, Kreuzer E, Reichart B. 2002. Is there an advantage in using homografts in patients with acute infective endocarditis of the aortic valve? J Heart Valve Dis 11:492-7.nHehrlein FW. 1971. Preservation problems of homologous and heterologous aortic valve grafts [in German]. Thoraxchir Vask Chir 19:394-5.nHörer J, Hanke T, Stierle U, et al. 2009. Homograft performance in children after the Ross operation. Ann Thorac Surg 88:609-15.nKilian E, Oberhoffer M, Gulbins H, Uhlig A, Kreuzer E, Reichart B. 2004. Ten years' experience in aortic valve replacement with homografts in 389 cases. J Heart Valve Dis 13:554-9.nLuk A, Butany J, Erlich SA, Henry J, David TE. 2007. Long-term morphological changes in a cryopreserved pulmonary valve homograft. Can J Cardiol 23:817-9.nLund O, Chandrasekaran V, Grocott-Mason R, et al. 1999. Primary aortic valve replacement with allografts over twenty-five years: valve-related and procedure-related determinants of outcome. J Thorac Cardiovasc Surg 117:77-90; discussion 90-1.nMair R, Harringer W, Wimmer-Greinecker G, et al. 1995. Aortic valve replacement with cryopreserved pulmonary allografts: five years' follow-up. Ann Thorac Surg 60:S185-8.nMirabet V, Carda C, Solves P, et al. 2008. Long-term storage in liquid nitrogen does not affect cell viability in cardiac valve allografts. Cryobiology 57:113-21.nNaegele H, Bohlmann M, Doring V, Kalmar P, Rodiger W. 2000. Results of aortic valve replacement with pulmonary and aortic homografts. J Heart Valve Dis 9:215-20; discussion 220-1.nNiwaya K, Elkins RC, Knott-Craig CJ, et al. 1999. Advantage of autograft and homograft valve replacement for complex aortic valve endocarditis. Ann Thorac Surg 67:1603-8.nO'Brien MF, Harrocks S, Stafford EG, et al. 2001. The homograft aortic valve: a 29-year, 99.3% follow up of 1,022 valve replacements. J Heart Valve Dis 10:334-44; discussion 335.nO'Brien MF, Stafford EG, Gardner M, et al. 1987. The viable cryopreserved allograft aortic valve. J Card Surg 2(suppl):153-67.nRiberi A, Caus T, Mesana T, et al. 1997. Aortic valve or root replacement with cryopreserved homograft for active infectious endocarditis. Cardiovasc Surg 5:579-83.nRoss DN. 1962. Homograft replacement of the aortic valve. Lancet 2:487-93.nSchütz A, Fischlein T, Haushofer M, et al. 1993. Homograft aus der eigenen Herzklappenbank. Z Transplantationsmedizin 5:194.nTweddell JS, Pelech AN, Frommelt PC, et al. 2000. Factors affecting longevity of homograft valves used in right ventricular outflow tract reconstruction for congenital heart disease. Circulation 102(suppl 3):III130-5.nYacoub M, Kittle F. 1970. Sterilization of homografts by antibiotic solutions. Circulation 41(suppl):II29-32.nYacoub M, Nasser RH, Lund O, et al. 1995. Fourteen-year experience with homovital homografts for aortic valve replacement. J Thorac Cardiovasc Surg 110:186-93.n
Published
How to Cite
Issue
Section
Author Disclosure & Copyright Transfer Agreement
In order to publish the original work of another person(s), The Heart Surgery Forum® must receive an acknowledgment of the Author Agreement and Copyright Transfer Statement transferring to Forum Multimedia Publishing, L.L.C., a subsidiary of Carden Jennings Publishing Co., Ltd. the exclusive rights to print and distribute the author(s) work in all media forms. Failure to check Copyright Transfer agreement box below will delay publication of the manuscript.
A current form follows:
The author(s) hereby transfer(s), assign(s), or otherwise convey(s) all copyright ownership of the manuscript submitted to Forum Multimedia Publishing, LLC (Publisher). The copyright transfer covers the exclusive rights to reproduce and distribute the article and the material contained therein throughout the world in all languages and in all media of expression now known or later developed, including but not limited to reprints, photographic reproduction, microfilm, electronic data processing (including programming, storage, and transmission to other electronic data record(s), or any other reproductions of similar nature), and translations.
However, Publisher grants back to the author(s) the following:
- The right to make and distribute copies of all or part of this work for use of the author(s) in teaching;
- The right to use, after publication in The Heart Surgery Forum, all or part of the material from this work in a book by the author(s), or in a collection of work by the author(s);
- The royalty-free right to make copies of this work for internal distribution within the institution/company that employs the author(s) subject to the provisions below for a work-made-for-hire;
- The right to use figures and tables from this work, and up to 250 words of text, for any purpose;
- The right to make oral presentations of material from this work.
Publisher reserves the right to grant or refuse permission to third parties to republish all or part of the article or translations thereof. To republish, such third parties must obtain written permission from the Publisher. (This is in accordance with the Copyright Statute, United States Code, Title 17. Exception: If all authors were bona fide officers or employees of the U.S. Government at the time the paper was prepared, the work is a “work of the US Government” (prepared by an officer or employee of the US Government as part of official duties), and therefore is not subject to US copyright; such exception should be indicated on signature lines. If this work was prepared under US Government contract or grant, the US Government may reproduce, royalty-free, all or portions of this work and may authorize others to do so, for official US Government purposes only, if the US Government contract or grant so requires.
I have participated in the conception and design of this work and in the writing of the manuscript and take public responsibility for it. Neither this manuscript nor one with substantially similar content under my authorship has been published, has been submitted for publication elsewhere, or will be submitted for publication elsewhere while under consideration by The Heart Surgery Forum, except as described in an attachment. I have reviewed this manuscript (original version) and approve its submission. If I am listed above as corresponding author, I will provide all authors with information regarding this manuscript and will obtain their approval before submitting any revision. I attest to the validity, accuracy, and legitimacy of the content of the manuscript and understand that Publisher assumes no responsibility for the validity, accuracy, and legitimacy of its content. I warrant that this manuscript is original with me and that I have full power to make this Agreement. I warrant that it contains no matter that is libelous or otherwise unlawful or that invades individual privacy or infringes any copyright or other proprietary right. I agree to indemnify and hold Publisher harmless of and from any claim made against Publisher that relates to or arises out of the publication of the manuscript and agree that this indemnification shall include payment of all costs and expenses relating to the defense of any such claim, including all reasonable attorney’s fees.
I warrant that I have no financial interest in the drugs, devices, or procedures described in the manuscript (except as disclosed in the attached statement).
I state that the institutional Human Subjects Committee and/or the Ethics Committee approved the clinical protocol reported in this manuscript for the use of experimental techniques, drugs, or devices in human subjects and appropriate informed consent documents were utilized.
Furthermore, I state that any and all animals used for experimental purposes received humane care in USDA registered facilities in compliance with the “Principles of Laboratory Animal Care” formulated by the National Society for Medical Research and the “Guide for the Care and Use of Laboratory Animals” prepared by the Institute of Laboratory Animal Resources and published by the National Institutes of Health (NIH Publication No. 85-23, revised 1985).