Totally Endoscopic Robotic Atrial Septal Defect Repair on the Beating Heart

Authors

  • Changqing Gao
  • Ming Yang
  • Gang Wang
  • Jiali Wang
  • Cangsong Xiao
  • Yang Wu
  • Jiachun Li

DOI:

https://doi.org/10.1532/HSF98.20091175

Abstract

Background: Atrial septal defect (ASD) repairs have successfully been performed on the arrested heart with the da Vinci S Surgical System (Intuitive Surgical). This study assessed the feasibility, safety, and efficacy of the use of the da Vinci S Surgical System for on-pump ASD repairs on the beating heart without cross-clamping the aorta.

Methods: This prospective study included 24 consecutive patients who underwent ASD repair surgery between June 2008 and June 2009. All of the procedures were completed with the da Vinci S robot via 3 port incisions in the right chest and a 1.5-cm working port. The operations were carried out on the beating heart with mild hypothermic cardiopulmonary bypass (CPB) without cross-clamping the aorta. Venting the heart from the working port provided adequate visualization of the operative field.

Results: All patients underwent complete repairs. Fourteen patients underwent ASD closure with a fresh autogenous pericardial patch, and 10 patients underwent direct ASD closure. Concomitant surgery was required in 4 patients. The mean (±SEM) CPB time was 65.6 ± 17.7 minutes, and the mean operative time was 98.5 ± 19.3 minutes. No patient required transfusion of red blood cells. The length of patient stay in the intensive care unit was 0.5 to 1.0 days. The length of hospital stay was 4 to 5 days. Follow-up transthoracic echocardiography evaluations showed no residual atrial septal leakage. There were no operative deaths, strokes, or other complications. All of the patients were discharged.

Conclusions: We have shown that use of the da Vinci S Surgical System to perform on-pump ASD repairs on the beating heart without cross-clamping the aorta is feasible, safe, and effective.

References

Blauth Cl, Cosgrove DM, Webb BW, et al. 1992. Atheroembolism from the ascending aorta. An emerging problem in cardiac surgery. J Thorac Cardiovasc Surg 103:1104-11.nDavila-Roman VG, Barzilai B, Wareing TH, et al. 1994. Atherosclerosis of the ascending aorta. Prevalence and role as an independent predictor of cerebrovascular events in cardiac patients. Stroke 25:2010-6.nGao C. 2008. Totally robotic internal mammary artery harvest and beating heart coronary artery bypass [abstract]. Heart Surg Forum 11(suppl 1):S279. Abstract OP-1167.nGao C, Yang M, Wang G, et al. 2007. Robotic atrial septal defect repair: preliminary experience with da Vinci S system. Chin J Thorac Cardiovasc Surg 23:298-300.nGao C, Yang M, Wang G, et al. 2007. Totally robotic internal mammary artery harvest and beating heart coronary artery bypass. Chin J Surg 45:1414-6.nGao C, Yang M, Wang G, et al. 2007. The first 4 patients undergoing totally robotic heart surgery without sternotomy. Chin J Thorac Cardiovasc Surg 23:19-21.nGao C, Yang M, Wang G, Wang J. 2008. Totally robotic resection of myxoma and atrial septal defect repair. Interact Cardiovasc Thorac Surg 7:947-50.nGao C, Yang M, Wu Y, et al. 2009. Hybrid coronary revascularization by endoscopic robotic coronary artery bypass grafting on beating heart and stent placement. Ann Thorac Surg 87:737-41.nMatsumoto Y, Watanabe G, Endo M, Sasaki H, Kasashima F, Kosugi I. 2002. Efficacy and safety of on-pump beating heart surgery for valvular disease. Ann Thorac Surg 74:678-83.nMiyairi T, Matsumoto J, Tanaka K, Mizuno A. 1996. Intraoperative assessment of functioning mitral valve. Ann Thorac Surg 61:743-5.nMizuno T, Toyama M, Tabuchi N, et al. 2000. Thickened intima of the aortic arch is a risk factor for stroke with coronary artery bypass grafting. Ann Thorac Surg 70:1565-70.nMurkin JM. 1999. Etiology and incidence of brain dysfunction after cardiac surgery. J Cardiothorac Vasc Anesth 13(suppl 1):12-7.nSalerno TA, Panos AL, Tian G, Deslauriers R, Calcaterra D, Ricci M. 2007. Surgery for cardiac valves and aortic root without cardioplegic arrest ("beating heart"): experience with a new method of myocardial perfusion. J Card Surg 22:459-64.nThompson MJ, Behranwala A, Campanella C, Walker WS, Cameron EW. 2003. Immediate and long-term results of mitral prosthetic replacement using a right thoracotomy beating heart technique. Eur J Cardiothorac Surg 24:47-51.n

Published

2010-06-09

How to Cite

Gao, C., Yang, M., Wang, G., Wang, J., Xiao, C., Wu, Y., & Li, J. (2010). Totally Endoscopic Robotic Atrial Septal Defect Repair on the Beating Heart. The Heart Surgery Forum, 13(3), E155-E158. https://doi.org/10.1532/HSF98.20091175

Issue

Section

Article