Endovascular Reconstruction from Aortic Valve to Aortic Arch Using One-Piece Valved-Fenestrated Stent Graft with a Branch: A Proof-of-Concept Study

Authors

  • Tao Li Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Xianhao Bao Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Jiaxuan Feng Department of Vascular Surgery, Changhai Hospital, Navy Medical University, Shanghai, China
  • Zhenjiang Li Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Junjun Liu Department of Vascular Surgery, The Affiliated Hospital of Qingdao University, Qingdao, Shandong, China
  • Yuxi Zhao Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Mingwei Wu Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Zhaoxiang Zeng Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Rui Feng Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Jian Zhou Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China
  • Zaiping Jing Endovascular Diagnosis and Treatment Center for Heart Valvular Diseases, and Endovascular Diagnosis and Treatment Center for Aortic Dissection, Changhai Hospital, Navy Medical University, Shanghai, China

DOI:

https://doi.org/10.1532/hsf.2585

Keywords:

endovascular repair; aortic root lesion; stent-graft

Abstract

Objective: To explore the feasibility of endovascular reconstruction of aortic root including aortic valve, sinus of valsalva and ascending aorta by a single stent-graft, a novel valved stent-graft with two fenestrae for preserving the coronary arteries was designed and performed in-vitro on a pig heart based circulation simulating system.

Methods: Pig hearts were harvested from 30 healthy adult female pigs weighing between 60-65 kilograms. Before sacrifice, all the pigs received aortic computed tomography angiography (CTA) examinations and morphologic parameters of aortic root were measured. Then we customized the valved stent-grafts according to the CTA measurements. After the pig heart was fixed on the special platform according to the original orientation and connected to the circulation system, the stent graft was delivered through transapical access and covered the segment from aortic annulus to proximal part of aortic arch under DSA (digital subtraction angiography) guidance. Then changes of coronary flow before and after the procedure and fenestration alignment with coronary ostia were analyzed.

Results: The overall technical success rate was 100%. The valve functions tested by ultrasound were in good condition under 120 mmHg circulation pressure. The flow of left coronary artery (LCA) did not decrease, but increased after the stent-graft implantation (340 ± 2.06 mL/min versus 288 ± 5.29 mL/min, P < .05). Similarly, the flow of right coronary artery (RCA) also increased (392 ± 9.17 mL/min versus 348 ± 8.01 mL/min; P < .05). The final angiography confirmed that both coronary arteries were patent. When generally observed from outer wall of valsalva sinus, both RCA and LCA orifices were aligned with the fenestrae. In 4 cases, the autologous valve leaflets blocked nearly 20% of the LCA fenestra’s area, but the flow did not significantly decrease in these cases.

Conclusion: Stimulated on a pig heart-based circulation simulation system, the one-piece valved-fenestrated stent graft with a branch could be delivered via the transapical access and deployed accurately, which achieved endovascular reconstruction of aortic valve, sinus of valsalva and ascending aorta while preserving the coronary artery perfusion by fenestrations. More in-vivo experiments on animal models are mandatory to further verify its efficacy and safety.

References

Bonow RO, Leon MB, Doshi D, Moat N. 2016. Management strategies and future challenges for aortic valve disease. Lancet 387:1312-23.

Chava S, Gentchos G, Abernethy A, Leavitt B, Terrien E, Dauerman HL. 2017. Routine CT angiography to detect severe coronary artery disease prior to transcatheter aortic valve replacement. J Thromb Thrombolysis 44:154-60.

Eltchaninoff H, Durand E, Avinee G, et al. 2018. Assessment of structural valve deterioration of transcatheter aortic bioprosthetic balloon-expandable valves using the new European consensus definition. EuroIntervention 14:e264-71.

Feuchtner GM, Schachner T, Friedrich G, Antretter H, Bonatti J, Zur ND. 2005. Acute aortic dissection with coronary ostium involvement and aortic valve regurgitation: three-dimensional visualization with multislice computed tomography. J Thorac Cardiovasc Surg 130:587-8.

Gulyaeva AS, Roshchevskaya IM. 2012. Morphology of moderator bands (septomarginal trabecula) in porcine heart ventricles. Anat Histol Embryol 41:326-32.

Mangieri A, Jabbour RJ, Montalto C, et al. 2017. Single-antiplatelet therapy in patients with contraindication to dual-antiplatelet therapy after transcatheter aortic valve implantation. Am J Cardiol 119:1088-93.

Smith CR. 2018. Emergency cardiac surgery following TAVI: implications for the future. Eur Heart J 39:685-6.

Wang C, Lachat M, Regar E, von Segesser LK, Maisano F, Ferrari E. 2018. Suitability of the porcine aortic model for transcatheter aortic root repair. Interact Cardiov Thorac Surg 26:1002-8.

Wipper S, Lohrenz C, Ahlbrecht O, et al. 2015. Transcardiac endograft delivery for endovascular treatment of the ascending aorta: a feasibility study in pigs. J Endovasc Ther 22:375-84.

Published

2019-09-16

How to Cite

Li, T., Bao, X., Feng, J., Li, Z., Liu, J., Zhao, Y., Wu, M., Zeng, Z., Feng, R., Zhou, J., & Jing, Z. (2019). Endovascular Reconstruction from Aortic Valve to Aortic Arch Using One-Piece Valved-Fenestrated Stent Graft with a Branch: A Proof-of-Concept Study. The Heart Surgery Forum, 22(5), E380-E384. https://doi.org/10.1532/hsf.2585

Issue

Section

Article