Right Ventricular Perforation by a Defibrillator Lead Migrating to the Left Breast

Authors

  • Jerneja Tasic Department of Cardiology, University Medical Center Ljubljana, Ljubljana, Slovenia
  • Rok Zbacnik Department of Clinical Institute of Radiology, University Medical Center Ljubljana, Ljubljana, Slovenia
  • Igor Zupan Department of Cardiology, University Medical Center Ljubljana, Ljubljana, Slovenia
  • Jus Ksela Department of Cardiovascular Surgery, University Medical Center Ljubljana

DOI:

https://doi.org/10.1532/hsf.1934

Keywords:

chest pain, defibrillator lead, heart perforation, thoracic wall perforation

Abstract

Cardiac perforation after an ICD implantation is a rare complication, with a reported incidence between 0.6-5.2%. Its manifestation might be acute, subacute, or delayed, with an acute perforation occurring within the first 24 hours after implantation, frequently accompanied by severe clinical signs, while subacute and delayed perforations have a more benign progression. Here, we report a case of a 69-year old patient with an acute right ventricular perforation by a defibrillator lead migrating all the way through the pericardium and thoracic wall into the left breast, with an unusually mild and benign clinical course, delaying prompt diagnosis and postponing subsequent surgical treatment. Heart perforation with a defibrillator electrode is a rare but dangerous complication, which may lead to pacing failure, cardiac tamponade, cardiogenic shock, and even death. Even with a benign clinical course, one must think of cardiac wall perforation at any time after device implantation, and a contrast enhanced computer tomography (CTA) must be performed if perforation is suspected. At re-implantation, the lead should be located at a different anatomical position within the RV, preferably at the interventricular septal site, and manipulation of the injury site within the RV avoided.

Author Biography

Jus Ksela, Department of Cardiovascular Surgery, University Medical Center Ljubljana

Department of Cardiovascular Surgery

Stuff surgeon

MD, PhD

Assistant Professor of surgery, Ljubljana Medical School

References

Banaszewski M, Stępińska J. 2012. Right heart perforation by pacemaker leads. Arch Med Sci 8:11-13.

Hsu JS, Varosy PD, Bao H, et al. 2013. Cardiac perforation from implantable cardioverter-defibrillatorlead placement. Insights from the National cardiovascular data registry. Circ Cardiovasc Qual Outcomes 6:582-590.

Khan MN, Joseph G, Khaykin Y, Ziada KM, Wilkoff BL. 2005. Delayed lead perforation: a disturbing trend. Pacing Clin Electrophysiol 28:251-3.

Kirkfeldt RE, Johansen JB, Nohr EA, et al. 2011. Risk factors for lead complications in cardiac pacing: A population-based cohort study of 28,860 Danish patients. Heart Rhythm 8:1622-1628.

Seil O. Cardiac Perforation Associated with a Pacemaker or ICD Lead, Modern Pacemakers - Present and Future, Prof. Mithilesh R Das (Ed.), 2011. ISBN: 978-953-307-214-2.

Sterlinski M, Przybylski A, Maciag, et al. 2009. Subacute cardiac perforations associated with active fixation leads. Europace 11: 206-212.

Published

2018-03-05

How to Cite

Tasic, J., Zbacnik, R., Zupan, I., & Ksela, J. (2018). Right Ventricular Perforation by a Defibrillator Lead Migrating to the Left Breast. The Heart Surgery Forum, 21(2), E087-E089. https://doi.org/10.1532/hsf.1934

Issue

Section

Article