Perfusion Flow Assessment of Coronary Shunt during Off-Pump Coronary Artery Bypass Grafting

Authors

  • Hirokuni Arai
  • Akira Kozakai
  • Susumu Manabe
  • Satoru Kawaguchi
  • Masato Shimizu
  • Koso Egi
  • Noriyuki Tabuchi
  • Hiroyuki Tanaka
  • Makoto Sunamori

DOI:

https://doi.org/10.1532/HSF98.200371320

Abstract

Background: Coronary shunts are widely used to prevent myocardial ischemia during off-pump coronary artery bypass graft (OPCAB) procedures. Although clinical effectiveness has been reported, actual perfusion flow has not been well assessed. The purpose of this study was to evaluate actual shunt flow and its pattern during passive coronary perfusion in clinical OPCAB.

Methods: In 15 OPCAB cases, the coronary perfusion flow of the external shunt (1.7 or 2.0 mm) during anastomosis and the free flow of the shunt were measured with an in-line electromagnetic or ultrasonic flow probe. The perfused coronary blood vessel was either the left anterior descending coronary artery or the right coronary artery. The inflow vessel of the external shunt was either the femoral artery (FA) or the ascending aorta (AA).

Results: Free flow values of a 1.7-mm FA shunt, 1.7-mm AA shunt, and 2.0-mm FA shunt were 34 ± 7, 39 ± 3, and 44 ± 7 mL/min. Perfusion flows were 13 ± 4, 14 ± 3, and 22 ± 4 mL/min, respectively. Perfusion flow was significantly lower than free flow and correlated well with coronary resistance. Although inflow site did not influence net perfusion flow, diastolic/systolic flow fraction ratio was significantly greater when the shunt was perfused from the FA.

Conclusions: External shunt from FA would provide limited but effective perfusion flow with a physiological pattern, which is passively regulated by coronary resistance.

References

Borges MF Spohn PK, Coulson AS. 1997. Arrhythmia/ischemia management during minimally invasive cardiac operation. Ann Thorac Surg 64:843-4.nDapunt OE, Raji MR, Jaschkeit S, et al. 1999. Intracoronary shunt insertion prevents myocardial stunning in a juvenile porcine MIDCAB model absent of coronary artery disease. Eur J Cardiothorac Surg 15:173-9.nDijk D, Nierich AP Jansen EWL, et al. 2001. Early outcome after off-pump versus on-pump coronary bypass surgery: results from a randomized study. Circulation 104:1761-6.nGüler M, Kirali K, Toker ME, et al. 2001. Different CABG methods in patients with chronic obstructive pulmonary disease. Ann Thorac Surg 71:152-7.nLee JD, Lee SJ, Tsushima WT et al. 2003. Benefits of off-pump bypass on neurologic and clinical morbidity: a prospective randomized trial. Ann Thorac Surg 76:18-26.nLevinson MM, Fooks GS. 1995. Coronary grafting using a temporary intraluminal shunt instead of heart-lung bypass. Ann Thorac Surg 60:1800-1.nLucchetti V Capasso F Caputo M, et al. 1999. Intracoronary shunt prevents left ventricular function impairment during beating heart coronary revascularization. Eur J Cardiothorac Surg 15:255-9.nMalkowski MJ, Kramer CM, Parvizi ST et al. 1998. Transient ischemia does not limit subsequent ischemic regional dysfunction in humans: a transesophageal echocardiographic study during minimally invasive coronary artery bypass surgery. J Am Coll Cardiol 31:1035-9.nMenon AK, Albes JM, Oberhoff M, Karsch KR, Ziemer G. 2002. Occlusion versus shunting during MIDCAB: effects on left ventricular function and quality of anastomosis. Ann Thorac Surg 73:1418-23.nOhtuka T Motomura N, Suematsu Y, Takamoto S. 2002. A simple distal coronary perfusion technique in off-pump coronary artery bypass. Eur J Cardiothorac Surg 22:443-4.nRivetti LA, Gandra SMA. 1997. Initial experience using an intraluminal shunt during revascularization of the beating heart. Ann Thorac Surg 63:1742-7.nRivetti LA, Gandra SMA. 1998. An intraluminal shunt for off-pump coronary artery bypass grafting: report of 501 consecutive cases and review of the technique. Heart Surg Forum 1:30-6.nYeatman M, Caputo M, Narayan P et al. 2002. Intracoronary shunts reduce transient intraoperative myocardial dysfunction during off-pump coronary operations. Ann Thorac Surg 73:1411-7.nAarnheim EEHL, Nierich AP Jansen EWL. 1999. When and how to shunt coronary circulation in off-pump coronary artery bypass. Eur J Cardiothorac Surg 16:S2-6.nArai H, Yoshida T Izumi H, Sunamori M. 2000. External shunt for off-pump coronary artery bypass grafting: distal coronary perfusion catheter. Ann Thorac Surg 70:681-2.nAscione R, Williams S, Lloyd CT Sundaramoorthi T Pitsis AA, Angelini GD. 2001. Reduced postoperative blood loss and transfusion requirement after beating-heart coronary operations: a prospective randomized study. J Thorac Cardiovasc Surg 121:689-96.nBittner HB, Savitt MA. 2002. Off-pump coronary artery bypass grafting decreases morbidity and mortality in a selected group of high-risk patients. Ann Thorac Surg 74:115-8.n